RCDD · Question #134
RCDD Question #134: Real Exam Question with Answer & Explanation
The correct answer is B. Oversized neutral and grounding conductors. Switching power supplies are non-linear loads that produce harmonic currents, which accumulate on the neutral conductor rather than canceling out, requiring oversized neutral and grounding conductors to prevent overheating.
Question
Options
- AIsolated grounding
- BOversized neutral and grounding conductors
- CSurge protection on power supply
- DA 3-phase feeder supply separate from the building supply
- EPower filtering at the service panel
Explanation
Switching power supplies are non-linear loads that produce harmonic currents, which accumulate on the neutral conductor rather than canceling out, requiring oversized neutral and grounding conductors to prevent overheating.
Common mistakes.
- A. Isolated grounding reduces noise coupling and ground-loop interference but does not address the overcurrent hazard on neutral conductors caused by harmonic accumulation.
- C. Surge protection on power supplies mitigates transient voltage spikes; it does not reduce or handle sustained harmonic current loading on the neutral conductor.
- D. A separate 3-phase feeder isolates the ER load from the building distribution but does not resolve the internal harmonic current buildup on the neutral conductor within the ER.
- E. Power filtering at the service panel can attenuate incoming harmonics but is not the standard primary countermeasure for harmonic-related neutral overloading inside an equipment room.
Concept tested. Neutral conductor oversizing for harmonic current from switching power supplies
Reference. https://www.nfpa.org/codes-and-standards/nfpa-70-standard-for-the-national-electrical-code